Hemodynamics Management

Public·282 members


This screenshot you provided shows the monitor of an Impella CP which appears to be positioned in the aorta instead of the correct position within the left ventricle (LV).

  • Ao Pressure: 126/73 mmHg (Mean 93 mmHg), indicating the device is picking up aortic pressure signals.

  • LV Pressure: There is a loss of clear pressure signals from the left ventricle, which would normally be used to monitor proper positioning.

  • The low fluctuation of the motor current, (with readings showing a current of 399/350 mA), suggests that there is minimal pressure gradient across the two pressure sensors of the Impella device. These sensors normally detect pressure differences between the aorta and the left ventricle.


The motor current stability (almost flat) suggests that the Impella is likely not in the left ventricle (as intended) but rather in the aorta. Since both sensors are exposed to aortic pressure, there is little to no pressure gradient detected…


Noor Ali Shah
Noor Ali Shah
Sep 21, 2024

This is totally new procedure for me.

Jazak Allah khair.


23 years old male with mental status changes, tachycardia and suppressed TSH level.

Based on the above echocardiogram his ventricular function is:

  • 0%Normal

  • 0%Mildly decreaded

  • 0%Moderately decreased

  • 0%Severely decreased


Dr.Yasser Alwali
Dr.Yasser Alwali
May 09, 2024

😞😞😞

Post cardiac arrest, CPR 15 min , ROSC achieved

Ibrahim Ameen
Ibrahim Ameen
Mar 04, 2024

I think the proper is flipped

The large akentic chamber in the left has thick wall similar to LV


VExUS for evaluation of venous congestion in a patient with mixed cardiogenic and distributive shock, acute respiratory failure, bibasilar infiltrates, and acute kidney injury.


Step 1: IVC diameter is 2.5 cm


Step 2: hepatic vein doppler assessment showing reversal of the systolic wave.


Step 3: Portal vein doppler assessment showing more than 50% variability


Ibrahim Ameen
Ibrahim Ameen
Nov 25, 2023

Very beautiful demonstration

It helps to suggest offloading rather than fluid resuscitation

It requires stady hands and minimal pt movement (abd wise) in order to catch the image in PW mode

If my memory serves me right, there were only limited data from multiple case report

No strong data to suggest outcome

Yet a useful tool to add on for the management

If we have research, i think mortality outcome, Cr trend, O2 requirements, vasopressors doses would be excellent markers for monitoring

Thank you Dr Mazen for your amazing continuous contribution to enriching our knowledge



ECLS-SHOCK trial investigated the effect of early extracorporeal life support (ECLS) on mortality in patients with heart attack-induced cardiogenic shock who were planned for early revascularization. Patients were assigned to ECLS plus medical treatment or medical treatment alone. After 30 days, mortality rates were similar between groups (ECLS: 47.8%, Control: 49.0%). ECLS patients had more bleeding (23.4% vs. 9.6%) and vascular complications (11.0% vs. 3.8%). ECLS didn't lower 30-day mortality in patients with infarct-related cardiogenic shock.


Extracorporeal Life Support in Infarct-Related Cardiogenic Shock | NEJM